LOL.
Matt.
96% CIS Lunar has a 2.66 j/l WOB at 300 feet and 75 RMV
66% AP Diving Inspiration has a 1.84 j/l WOB at 300 feet and 75 RMV
67% MK16 has a 1.85 j/l at 300 feet and 75 RMV
75% MK16 Mod 2 has a 2.08 j/l at 300 feet and 75 RMV
63% Stealth MOD has a 1.71 j/l at 300 feet and 75 RMV
66% Viper E Stealth MOD has a 1.82 j/l at 300 feet and 75 RMV
62% JJ has a 1.71J/l WOB at 100m and 75 RMV (11/65)
At 75 RMV the Inspiration is 1.84 j/l at 100.3m or 329ft
At 89.9m (295ft) it's 1.75 j/l
By interpolating, at 300 ft = 1.76 to 1.77 j/l
(All tests done at 4.3 C, all figures ATP, no BTPS adjustment. )
To put it into perspective, the limit is 2.75 j/l but with BTPS conversion. To find the 75 lpm BTPS level I'd have to test at 62.5 lpm and then convert both to BTPS and interpolate between the two. 75 RMV (ATP) was 82 (BTPS) from memory at 4.3C and 100.3m.
In 2002 when the NEDU test was done, the WOB limit was the one for OC regulators, 3.0 j/l. They tested one over this and one under, giving their average of 2.98. having seen much lower figures at QinetiQ I know there's an error in their testing.
All test houses are capable of making mistakes and I have challenged NEDU on their figures. We'll see whether they can be bothered to re-test
Since 2004, the Inspirations have been inside the EN14143:2003 limits and this has been proven by extra audits and external testing.
But, as I pointed out in my first post... A p*ssing contest on WOB is a waste of time, you need to take into account elastance and hydrostatic imbalance as well.
Humans are really good at feeling hydrostatic imbalance and often confuse that with work of breathing, elastance takes them a little longer to pick up but the pain arrives eventually, WOB or resistive flow historically isn't so easy to feel...look at the regulators produced in the 70s and 80s...there were probably only a couple of models that would meet today's EN250, (3 j/l limit at 50m (air) at 62.5 RMV.)
It's taken years to get agreement on how to even measure elastance and hydrostatic imbalance between both sides of the Atlantic and I'm not convinced the Hydrostatic imbalance test method is where it needs to be yet.
Dan Warkander's proposal is to add all three together, by adding the % of max. for each value, how valid that is I don't know but it would certainly give divers a better comparison tool. All figures would have to be gleaned in all positions which as Dave said is less onerous when you have your own machine and is something for the future but to have all those tests done and externally verified would be mega expensive and I can't see that getting through the standards committee without a lot of proof as to it's worth.
While I am writing about the European Standards Committee, Gian's comments about the makeup of the committee is wrong. As far as I am aware, explanations have been given to him as to why his claims for defective standards have been rejected. He claims commercial interest drove those decisions but my understanding is that the make up of the committee was provided which would make it clear that the make up is a balanced group comprised of physiologists, test houses, user group representatives as well as manufacturers and training organisations.
My understanding is that collectively and unanimously they rejected his claims.