Dave,
I think we have to accept that we have an infra-red CO2 sensor that will accurately measure CO2 in a sample of gas presented to it in the rebreather environment. Iain Middlebrook has confirmed this. That being the case, you have to do one of three things in your project:
Either:
Find a way of putting the sensor in the mouthpiece. This would be the ideal option because you would be able to measure inpired CO2 during inspiration, and the end tidal CO2 during expiration. The presence of CO2 during inspiration means your scrubber is failing, and the end tidal CO2 will tell you if the diver is hypercapnic from any cause (CO2 rebreathing or CO2 retention). Indeed, with the combination of inspired and end tidal readings you can diagnose the cause of hypercapnia and formulate a sensible response. If the end tidal CO2 is high (ie the diver is hypercapnic) and there is CO2 in the inspired gas then the probable cause is scrubber failure and bailing out is the correct response. If the end tidal CO2 is high and the there is no CO2 in the inspired gas then the probable cause is CO2 retention and reducing both exercise and your work of breathing is the goal. Bailing out off a good loop may not be the best course of action in this setting.
Why has this not been done? Size of the sensor probably. You really would need a gag strap to hold your mouthpiece in place! Arne Sieber is working on solid state CO2 sensing technology that will be small enough to put in a mouthpiece but this is still a way off I think.
Or:
Find a way to sample gas from the mouthpiece to a sensor in the back pack. In our anaesthesia circuits this is achieved by a pump that draws gas at about 200ml/min through a fine capillary tube to the sensor unit. This is effectively the same as measuring CO2 at the mouth and has all the advantages outlined above.
Why has this not been done? Power required for the pump probably. I have always harboured a suspicion that a clever engineer might be able to devise a way of doing this using the divers respiratory effort to power a pump device but this is way outside my area of expertise. I don't think it would be easy.
Or:
Stick with Alex's original method using direct measurement at the end of the hose whilst acknowledging its limitations and potential inaccuracies. This is obviously the least desirable option because of the potential inaccuracies and the constraints they place on interpretation of the numbers. However, I must acknowledge the fact that in our published simulation of his technique it worked fairly well at higher tidal volumes (breath sizes). Alex appeared to change it after the original debate (leaving the sensor where it was, but applying a complex mathematical algorithm to derive an end tidal CO2 from the reading). To be honest, this seems even less likely to work. There has been a lot of talk about it having been tested but no data has been presented and quite frankly, based on the sticking oxygen buttons, flying off lids and leaking components etc on the oxygen units, I find myself doubting the credibility of the test program.
So there it is. If I were a handy person like yourself, the first thing I would do is have a look at all the sensors that appear capable of doing the job (Iain has strong opinions on this and can advise you), check out their dimensions, and evaluate the plausibility of incorporation in a mouthpiece.
Simon M
I think we have to accept that we have an infra-red CO2 sensor that will accurately measure CO2 in a sample of gas presented to it in the rebreather environment. Iain Middlebrook has confirmed this. That being the case, you have to do one of three things in your project:
Either:
Find a way of putting the sensor in the mouthpiece. This would be the ideal option because you would be able to measure inpired CO2 during inspiration, and the end tidal CO2 during expiration. The presence of CO2 during inspiration means your scrubber is failing, and the end tidal CO2 will tell you if the diver is hypercapnic from any cause (CO2 rebreathing or CO2 retention). Indeed, with the combination of inspired and end tidal readings you can diagnose the cause of hypercapnia and formulate a sensible response. If the end tidal CO2 is high (ie the diver is hypercapnic) and there is CO2 in the inspired gas then the probable cause is scrubber failure and bailing out is the correct response. If the end tidal CO2 is high and the there is no CO2 in the inspired gas then the probable cause is CO2 retention and reducing both exercise and your work of breathing is the goal. Bailing out off a good loop may not be the best course of action in this setting.
Why has this not been done? Size of the sensor probably. You really would need a gag strap to hold your mouthpiece in place! Arne Sieber is working on solid state CO2 sensing technology that will be small enough to put in a mouthpiece but this is still a way off I think.
Or:
Find a way to sample gas from the mouthpiece to a sensor in the back pack. In our anaesthesia circuits this is achieved by a pump that draws gas at about 200ml/min through a fine capillary tube to the sensor unit. This is effectively the same as measuring CO2 at the mouth and has all the advantages outlined above.
Why has this not been done? Power required for the pump probably. I have always harboured a suspicion that a clever engineer might be able to devise a way of doing this using the divers respiratory effort to power a pump device but this is way outside my area of expertise. I don't think it would be easy.
Or:
Stick with Alex's original method using direct measurement at the end of the hose whilst acknowledging its limitations and potential inaccuracies. This is obviously the least desirable option because of the potential inaccuracies and the constraints they place on interpretation of the numbers. However, I must acknowledge the fact that in our published simulation of his technique it worked fairly well at higher tidal volumes (breath sizes). Alex appeared to change it after the original debate (leaving the sensor where it was, but applying a complex mathematical algorithm to derive an end tidal CO2 from the reading). To be honest, this seems even less likely to work. There has been a lot of talk about it having been tested but no data has been presented and quite frankly, based on the sticking oxygen buttons, flying off lids and leaking components etc on the oxygen units, I find myself doubting the credibility of the test program.
So there it is. If I were a handy person like yourself, the first thing I would do is have a look at all the sensors that appear capable of doing the job (Iain has strong opinions on this and can advise you), check out their dimensions, and evaluate the plausibility of incorporation in a mouthpiece.
Simon M
Last edited: